© 2021 IEEE International Solid-State Circuits Conference
ISSCC Forum 1 of 78
ISSCC 2021 ForumsShuenn-Yuh Lee received the B.S. degree from the National Taiwan Ocean
University, Keelung, Taiwan, in 1988, and the M.S. and Ph.D. degrees from the
National Cheng Kung University, Tainan, Taiwan, in 1994 and 1999, respectively. He
is currently a Professor at the Department of Electrical Engineering, National Cheng
Kung University, Tainan, Taiwan. From 2013 to 2016, he serves as the Chairman of
IEEE Solid-State Circuits Society Tainan Chapter. From 2016 to 2017, he serves
as the Vice Chairman of IEEE Tainan Section. He is the Associate Editor of IEEE
Transaction on Biomedical Circuits and Systems from 2016-current. His present
research activities involve the design of analog and mixed-signal integrated
circuits, biomedical circuits and systems, low-power and low-voltage analog
circuits, and RF front-end integrated circuits for wireless communications.
© 2021 IEEE International Solid-State Circuits Conference
Cardiovascular Disease Detection, Analysis and Evaluation System-On-Chip and Platform
Shuenn-Yuh Lee
[email protected] Department/National Cheng Kung University
February 21, 2021
ISSCC Forum 2 of 78
ISSCC 2021 Forums
© 2021 IEEE International Solid-State Circuits Conference
Outline Introduction to ECG signal
Low-power wireless ECG acquisition and classification system for body sensor networks
A low-power bidirectional telemetry device with a cardiac microstimulator for implantable body sensor networks
Portable and wireless urine detection system and platform for prevention of cardiovascular disease
ConclusionISSCC Forum 3 of 78
© 2021 IEEE International Solid-State Circuits Conference
Motivation: Top Ten Causes of Death [1]
[1] https://www.who.int/news-room/fact-sheets/detail/the-top-10-causes-of-deathISSCC Forum 4 of 78
Unit: millions people
20002019
© 2021 IEEE International Solid-State Circuits Conference
Impulse Conduction & ECG
https://en.wikipedia.org/wiki/Heart
Sinus Node
AV Node
Bundle of His
Bundle Branches
Purkinje Fibers
Sinus Node
AV Node
Bundle of His Right Bundle Branch
Left Bundle Branch
Atrium
Ventricles
Purkinje Fibers
ISSCC Forum 5 of 78
© 2021 IEEE International Solid-State Circuits Conference
Waves of the ECG (Electrocardiogram)
https://en.wikipedia.org/wiki/ElectrocardiographyISSCC Forum 6 of 78
P wave: atrial depolarizationQRS wave: ventricular depolarization
T wave: ventricular repolarizationU wave: slow repolarization of papillary muscles
© 2021 IEEE International Solid-State Circuits Conference
ECG Patch Requirement
Monitoring time Classification Adjustment Feedback Representative
graph
Holter 24-48 hrs off line Y N
Zio Patch 7 days off line N NRefit Patch 8 hrs on line N
N
Leadtek(amor H1) 24 hrs N N N
(0.05~40HZ)
This work
>24 hrs6 months (off-line)
on line Y Y(0.05Hz ~ 150Hz)
ISSCC Forum 7 of 78
© 2021 IEEE International Solid-State Circuits Conference
Challenges Home-care system
Long-term usage
Low-power consumption
Convenient usage
Wireless transmission
High resolution
For diagnosing precisely
Diagnosis
For both user and the doctorISSCC Forum 8 of 78
© 2021 IEEE International Solid-State Circuits Conference
Interactive Intelligent Healthcare System Targets Low power, Anywhere, Anytime, Long-term monitoring
Body Sensor Network Near-body application Wearable device
Local Sensor Network Portable device (smartphone, notebook, PDA)
ISSCC Forum 9 of 78
© 2021 IEEE International Solid-State Circuits Conference
Outline Introduction to ECG signal
Low-power wireless ECG acquisition and classification system for body sensor networks
A low-power bidirectional telemetry device with a cardiac microstimulator for implantable body sensor networks
Portable and wireless urine detection system and platform for prevention of cardiovascular disease
ConclusionISSCC Forum 10 of 78
© 2021 IEEE International Solid-State Circuits Conference
Low-power Wireless ECG Acquisition Circuit and System for Body Sensor Networks Pro of SOC
Personal device
Cloud server
HW & SW
Special in our SOC
Self-developed SOC
Low cost, low power
High flexibility & reliabilityISSCC Forum 11 of 78
© 2021 IEEE International Solid-State Circuits Conference
Proposed System Overview
Bio-signal processor (BSP): Signal acquisition and digitization Transceiver: wireless transmission for 1 meter distance Digital-signal processor (DSP): Signal classification
ISSCC Forum 12 of 78
© 2021 IEEE International Solid-State Circuits Conference
Proposed System Overview
Simple BSP architecture with high resolution Transceiver with low power consumption Signal feature extraction and classification
ISSCC Forum 13 of 78
© 2021 IEEE International Solid-State Circuits Conference
Bio-Signal Processor Conventional BSP design (Analog front-end circuits)
Conventional design strategy Simple architecture with lower resolution Complex architecture with higher resolution Performance is always limited by filters (< 8-10 bits)
ISSCC Forum 14 of 78
© 2021 IEEE International Solid-State Circuits Conference
Proposed Bio-Signal Processor
Eliminate the low-frequency noises, ex: flicker noise, dc offset
Remove the filter circuit
Digitizing directly by the high-pass sigma delta modulator (HPSDM)ISSCC Forum 15 of 78
© 2021 IEEE International Solid-State Circuits Conference
Proposed Bio-Signal Processor (cont.) Advantages of the BSP: Simple architecture: prone to be integrated High performance: by applying SDM and eliminating the filter One bit stream output: Reduce parallel-to-serial circuits Against transmission error: reduce encoding/ encryption circuit
Specifications: Architecture CBCTA+HPSDMClose loop gain > 20dB
Resolution > 10bitsBandwidth 200Hz
Chopping frequency 25.6kHzSampling frequency 51.2kHz
OSR 128Output format modulated 1-bit
stream Power consumption As low as possible
ISSCC Forum 16 of 78
© 2021 IEEE International Solid-State Circuits Conference
High Pass Sigma Delta Modulator (of BSP)Spec. Value Unit Implementation ParametersENOB > 12 bit Architecture DTSDM
Oversampling Ratio 128 Hz/ Hz Modulator order 3 OrderBandwidth 200 Hz Quantizer 1 Bit
Sampling Freq. 51.2k HzTopology Feed-forward with local
feedback pathMax. Input -20 dBV
-1
-1z
1+z
-1
-1z
1+z-1
-1z
1+za1 2.5 g1 0.333
a2 1.25 g2 0.667
a3 1.25 g3 0.179b1 0.01
Coefficient table
ISSCC Forum 17 of 78
© 2021 IEEE International Solid-State Circuits Conference
High Pass Sigma Delta Modulator (of BSP) High-pass integrator [2]
[2] V. T. Nguyen, et. al., “VHDL-AMS behavioral modeling and simulation of high-pass delta-sigma modulator,” IEEE BMAS 2005)
ISSCC Forum 18 of 78
© 2021 IEEE International Solid-State Circuits Conference
High Pass Sigma Delta Modulator (of BSP) Circuit schematic & capacitor values
CS1 0.17pF CI1 0.51pF CH1 1.02pF CA1 0.10pF
CS2 0.10pF CI2 0.15pF CH2 0.3pF CA2 0.05pF
CS3 0.10pF CI3 0.56pF CH3 1.12pF CA3 0.05pF
CI4 0.04pF CB1K 2.71pF CB1 0.10pF
ISSCC Forum 19 of 78
© 2021 IEEE International Solid-State Circuits Conference
Simulation Results of the BSP One tone & two-tone test SNR > 72 dB SFDR > 72 dB
ISSCC Forum 20 of 78
© 2021 IEEE International Solid-State Circuits Conference
Specifications of the BSPGeneral
Technology TSMC 0.18 μm 1P6M processSupply voltage 1.2 V
CBCTAMid-band gain 20/ 24/ 28 dB
PSRR 73.8 dBCMRR 109.6 dB
Input refer noise 1.2 μVRMS (25.4–25.6 kHz)SFDR 76.34 dB (with 5,000 μV input)
Chopper frequency 25.6 kHzPower consumption 15 μW
HPSDMArchitecture High-pass, third-order, feed-forward SDM
SNR 72 dB (with 0.1 V input signal)SFDR 75.4 dB (with two-tone test)
Oversampling ratio 128Sampling frequency 51.2 kHzPower consumption 13.5 μW
Entire BSP circuitsPower consumption 28.7 μW
Resolution > 12 bits
ISSCC Forum 21 of 78
© 2021 IEEE International Solid-State Circuits Conference
Transceiver The demands of wireless transmission for the proposed BSNs Low power: for long-term usage
Short-range transmission: for 1-meter distance
Simple architecture: for system integration
Considerations of the specifications Safety is the must concerned issue on the BSNs
Frequency selection
Transmission power & distance
Data rate, power consumption & architectureISSCC Forum 22 of 78
© 2021 IEEE International Solid-State Circuits Conference
Transceiver (cont.) Consideration of the specifications Safety: SAR (Specific Absorption Rate) versus frequency [3] Frequency selection Transmission power & distance
[3] E. Cocherova, et. al. “Dependence of the RF field absorption on the human body dimensions,” International Conference Radioelektronika, 2009.
ISSCC Forum 23 of 78
© 2021 IEEE International Solid-State Circuits Conference
Transceiver (cont.) Consideration of the specifications Safety Frequency selection: ISM Band Transmission power & distance
Frequency range Center frequency Availability
13.553-13.567 MHz 13.560 MHz -26.957-27.283 MHz 27.120 MHz -
40.66-40.70 MHz 40.68 MHz -433.05-434.79MHz 433.92 MHz ITU Region I only
902-928 MHz 915 MHz ITU Region II only2.400-2.500 GHz 2.45 GHz -5.725-5.875 GHz 5.800 GHz -
Region 1, 2 & 3 [4]
[4] http://life.itu.int/radioclub/rr/frr.htm
ISSCC Forum 24 of 78
© 2021 IEEE International Solid-State Circuits Conference
Transceiver (cont.) Consideration of the specifications Transmission power & distance
Assume that Transmitter is mounted on the body : 1cm or 2cm (R) thickness transmitting power: -10 dBm (Pt), antenna gain: 0 dBi (Gt) Power density => 0.08 and 0.02 W/m2 => safety
Power density (W/m2)Controlled exposure(6 minutes average)
Uncontrolled exposure(6 minutes average)
IEEE [8] 80 50ICNIRP [9] 50 10
2 ( )4
t tPGPower Density SRπ
=
[5] IEEE Standard C95.1-1999,[6] ICNIRP safety guideline, April 1998.
ISSCC Forum 25 of 78
© 2021 IEEE International Solid-State Circuits Conference
Transceiver (cont.) Consideration of the specifications Data rate, Power consumption & architecture
Architecture Sensitivity Data rateCircuit
complexity
Power
consumption
Super
Heterodyne★★ ★★ ★★ ★★
Direct
Conversion★☆ ★☆ ☆☆ ★☆
Low IF ★☆ ★☆ ★☆ ★☆Super
Regenerative☆☆ ☆☆ ☆☆ ☆☆
ISSCC Forum 26 of 78
© 2021 IEEE International Solid-State Circuits Conference
Transceiver (cont.)
Architecture comparison Super-regenerative: Sensitivity ↓, Power ↓, Circuit complexity ↓
ISSCC Forum 27 of 78
© 2021 IEEE International Solid-State Circuits Conference
Super-regenerative Transceiver
Transmitter: VCO, buffer, modulator Receiver: LNA, VCO, current steering DAC, envelop detector, comparator
ISSCC Forum 28 of 78
© 2021 IEEE International Solid-State Circuits Conference
Transceiver (cont.) Specifications:
Architecture Super regenerative OOK transceiver
Transmission distance 1 meterTransmitter
Output power > -20dBmAntenna gain 0dBi
Output frequency 2.4GHz~ 2.48GHz (ISM band)Receiver
Sensitivity < -60dBmAntenna gain 0 dBi
Received frequency 2.4GHz – 2.48GHz
ISSCC Forum 29 of 78
© 2021 IEEE International Solid-State Circuits Conference
Digital Signal Processor
Two main functions:
Signal demodulation: shifting back the bio-signal
Signal analysis: using wavelet transform for ECG classificationISSCC Forum 30 of 78
© 2021 IEEE International Solid-State Circuits Conference
Digital Signal Processor- WT (cont.) Beat detection
R-wave
Beats per minute
Classification
Compare the results
From MCS
ISSCC Forum 31 of 78
© 2021 IEEE International Solid-State Circuits Conference
Digital Signal Processor (cont.)
Main function
DemodulationDecimation
Coefficient extraction by Wavelet
transformBeat detection
DecimatorDecimation rate 128
Resolution >10 bitsWavelet transformer
Resolution > 8bitsISSCC Forum 32 of 78
© 2021 IEEE International Solid-State Circuits Conference
Chip Implementation TSMC 0.18μm 1P6M standard process with supply voltage of 1.2V BSP +TX: 1.55 x 1.51 mm2
RX + decimator: 1.28 x 2.04 mm2
DSP (wavelet transform) : 1.57 x 1.57 mm2
BSP+TX RX+decimator DSP
ISSCC Forum 33 of 78
© 2021 IEEE International Solid-State Circuits Conference
Summary A wireless ECG acquisition and classification system is proposed BSN architecture High-performance : > 10 bits Low power consumption : body-end circuits operated> 100 days Wireless transmission : 1 meter distance ECG classification : storing 8 kinds of reference data Simplified architecture
3 chips implementation Body-end circuits: BSP + TX Receiving-end circuits: RX + decimator, Wavelet transformer
ISSCC Forum 34 of 78
© 2021 IEEE International Solid-State Circuits Conference
Applications (Replace Holter) Arrhythmia monitoring system
1.69 mm
1.73
mm
CCTA HPSDM
DSP
ISSCC Forum 36 of 78
© 2021 IEEE International Solid-State Circuits Conference
Ch1:V5
Ch2:V1
Ch3:Lead III
Dotted circles:Our device (ECG patch)
Method (Holter Medical Device)White(1-)
Blue(3-)Black(2-)
Red(1+)
Orange(3+)
Brown(2+)
Green(RL, GND)
ISSCC Forum 37 of 78
© 2021 IEEE International Solid-State Circuits Conference
Patients use two devices at the same time Check the relationship between our device and Holter
Self-developed APP
Method (Holter Medical Device)
Similar with Ch3 and the rhythm is almost the same. A PVC is classified by ECG Patch and Holter
ISSCC Forum 38 of 78
© 2021 IEEE International Solid-State Circuits Conference
VPB :ventricular premature beat (also called PVC) VT :ventricular tachycardia SVPB :supraventricular premature beats (also called APC) SVT :supraventricular tachycardia
Holter Format
ISSCC Forum 39 of 78
© 2021 IEEE International Solid-State Circuits Conference
No.200 in MIT-BIH arrhythmia database Rate: 60-100 bpm Regularity: irregular P waves: abnormal-not present on VPC PR interval: miss QRS duration: bizarre and >0.12 sec.
Premature Ventricular Contraction (PVC)
ISSCC Forum 40 of 78
© 2021 IEEE International Solid-State Circuits Conference
No.100 in MIT-BIH arrhythmia database Rate: 60-100 bpm Regularity: occasionally irregular P waves: normal PR interval: 0.12-0.20 sec QRS duration: 0.04-0.12 sec.
Atrial Premature Contraction (APC)
ISSCC Forum 41 of 78
© 2021 IEEE International Solid-State Circuits Conference
No.205 in MIT-BIH arrhythmia database Rate: >100 bpm Regularity: irregular P waves: not present on VPC (three or more) PR interval: miss QRS duration: 0.04-0.12 sec.
Ventricular Tachycardia (VT)
ISSCC Forum 42 of 78
© 2021 IEEE International Solid-State Circuits Conference
The Bluetooth connection is stable 24 hours data recorded Normal rhythm
Similar with Ch3 and the rhythm is the same
Fifth Patient in Tainan Hospital
ISSCC Forum 43 of 78
© 2021 IEEE International Solid-State Circuits Conference
An APC occurred Detected by Holter & our device
Noted as red circle
Fifth Patient in Tainan Hospital (cont.)
ISSCC Forum 44 of 78
© 2021 IEEE International Solid-State Circuits Conference
PVC pair occurred Noted as red circle
Fifth Patient in Tainan Hospital (cont.)
ISSCC Forum 45 of 78
© 2021 IEEE International Solid-State Circuits Conference
Summary A bio-signal acquisition and arrhythmia classification system is proposed
and implemented in TSMC 0.18 μm standard CMOS process. The AFE has 10 bits resolution digital output and its power consumption is
28.7 μW. A wavelet based QRS detection and arrhythmia classification method was
proposed. The accuracy in terms of sensitivity and positive predictionbased on the MIT-BIH arrhythmia database are 99.44% and 99.40%,respectively. The accuracy of arrhythmia classification is 95.83 %.
There are more than 100 patients are enrolled in the clinical trial. Thearrhythmias can be detected by Holter and also classified by our device.
Yutech Co., Ltd. (start-up company): Founded in May of 2019 at Tainan,Taiwan (https://www.yutechealth.com/index_en.html)
ISSCC Forum 46 of 78
© 2021 IEEE International Solid-State Circuits Conference
Outline Introduction to ECG signal
Low-power wireless ECG acquisition and classification system for body sensor networks
A low-power bidirectional telemetry device with a cardiac microstimulator for implantable body sensor networks
Portable and wireless urine detection system and platform for prevention of cardiovascular disease
ConclusionISSCC Forum 47 of 78
© 2021 IEEE International Solid-State Circuits Conference
Motivation Therapy by micro-stimulation
ISSCC Forum 48 of 78
© 2021 IEEE International Solid-State Circuits Conference
A low-power cardiac microstimulator for implantable body sensor networks
System overview
Circuitry design
Experimental results
Summary
ISSCC Forum 49 of 78
© 2021 IEEE International Solid-State Circuits Conference
System Overview Problem in conventional devices: large volume and high power
consumption Improved by four blocks: powering interface, digital circuitry, pacing
channel, sensing channel
ISSCC Forum 50 of 78
© 2021 IEEE International Solid-State Circuits Conference
Proposed Close-Loop Micro-Stimulator Powering interface: RF front-end, charging circuit, charge pump Digital circuitry: PSK demodulator, system controller Pacing channel: Micro-stimulator Sensing channel: Amplifier, filter, ADC, LSK modulator
ISSCC Forum 51 of 78
© 2021 IEEE International Solid-State Circuits Conference
A low-power cardiac microstimulator for implantable body sensor networks
System overview
Circuitry design
Experience results
Summary
ISSCC Forum 52 of 78
© 2021 IEEE International Solid-State Circuits Conference
Powering Interface Main features: rechargeable device, LDO regulator, high efficient
charge pump Small battery body Stable AC/DC converter and DC/DC converter
ISSCC Forum 53 of 78
© 2021 IEEE International Solid-State Circuits Conference
Digital Circuitry
ISSCC Forum 54 of 78
Provide programmable functions and enhance reliability of communication PLL based PSK demodulator System controller flow chart
© 2021 IEEE International Solid-State Circuits Conference
Pacing Channel
Atrial Strength-Duration Curve
ISSCC Forum 55 of 78
© 2021 IEEE International Solid-State Circuits Conference
A low-power cardiac microstimulator for implantable body sensor networks
System overview
Circuitry design
Experience results
Summary
ISSCC Forum 57 of 78
© 2021 IEEE International Solid-State Circuits Conference
Chip Microphotograph TSMC 0.35µm 2P4M 1.5×1.6mm2
Wireless telemetry pacemaker test setup
Coils radius: 12.5mm / 7.5mm
Distance : 10mm~20mm
ISSCC Forum 58 of 78
© 2021 IEEE International Solid-State Circuits Conference
Programmable Function Pulse amplitude, pulse duration, pulse frequency
ISSCC Forum 60 of 78
© 2021 IEEE International Solid-State Circuits Conference
Measurement- Close Loop Test
ISSCC Forum 61 of 78
© 2021 IEEE International Solid-State Circuits Conference
Measured SummaryGeneral
Technology TSMC 0.35µm 2P
Chip area 1.25×1.85mm2
Carrier frequency 256kHzPower Interface and Pacing Channel
Rectifier 2 V output with induced 2.9 Vpp
Regulator power 7 µW @ 1V
Charge Pump
power 10 μWoutput voltage
(@ 1 V input voltage) 3.2 V
pumping clock frequency 16 kHz
PowerManagement
(Supply Detector,Charge Detector,
and Charger)
power 20 μWcharging current 2 mA
Battery(6 mAh @ 1.2 Vsupply voltage)
2 x V6HR,(1000 times recharging)
D/A Controllerpower 42nW @ 1V
operation frequency 8 kHz
Pulse Generatorpower
(pacing period = 1s;pulse duration = 0.5ms)
15.5µW @ 5.8V
Digital Blocks
PSK Demodulator power 1.76µW @ 1Vsampling frequency 4MHz
System Controller power 260nW @ 1Voperation frequency 32kHz
ProgrammableRanges
stimulation frequency 0.5–31.25 Hzstimulus amplitude 0–3.2 V
stimulus duration 62.5 μs–1.94 ms
Monitoring Analog Front End @ 1.4V
Preamplifier
power 40nWInput referred noise 1.1µVrms
3-dB bandwidth 0-140HzDC gain 21/31dB
Biquad Low-Pass/Band-Pass Filter
power 126 nW(with buffer)
3-dB BandwidthLow-Pass 0–15 HzBand-Pass 15–80 Hz
Real-time Detection ADC
power 71nW(with SCAMP)
resolution 8bitssampling frequency 800Hz
sample-and-hold gain 2/5 times
ISSCC Forum 63 of 78
© 2021 IEEE International Solid-State Circuits Conference
Summary
Proposed a programmable implantable micro-stimulator
SoC with wireless telemetry
Smart powering management: reduce the device volume
Low power analog front end sensing channel: enhance the
battery life
Wireless telemetry: on-line monitor and programming
Programmable control: multi-function stimulationsISSCC Forum 64 of 78
© 2021 IEEE International Solid-State Circuits Conference
Outline Introduction to ECG signal
Low-power wireless ECG acquisition and classification system for body sensor networks
A low-power bidirectional telemetry device with a cardiac microstimulator for implantable body sensor networks
Portable and wireless urine detection system and platform for prevention of cardiovascular disease
ConclusionISSCC Forum 65 of 78
© 2021 IEEE International Solid-State Circuits Conference
Motivation
Around 15% people in 75years and older group sufferfrom valvular heart disease
Provides a telecare platform to monitor the risks of cardiovascular disease (CVD) and prevent CVD
ISSCC Forum 66 of 78
© 2021 IEEE International Solid-State Circuits Conference
Proposed Electrochemical Sensing System
ISSCC Forum 67 of 78
RDAC
VCM
Potentiostat
Current-sensing VCO-based SDM
RW2
MUX
C
W
Sel [1:0]W3W1
Controller +Decimator +
UART
MCU+BLE(CYBLE-022001) ADC Output
DAC Output
UART
Power Management
Bandgap Reference
3.3V/1.8V LDO
T18_109B Chip
© 2021 IEEE International Solid-State Circuits Conference
UACR sensor (Urine Albumin and Creatinine Sensors)
68 of 78
Creatinine sensor
SPCE|ABTS-CNT|Nafion® electrode
Albumin sensor
SPCE|Au|Ab-HSAimmunosensor• Sensitivity: 27.3 ± 0.5 µA cm-2 mM-1
• LOD: 11.0 µM• linear range: 16.5mM ~26.84mM
ISSCC Forum
© 2021 IEEE International Solid-State Circuits Conference
Urine Detection Circuit Right: Conventional architecture of potentiostat with the TIA-topology
Left: Proposed readout circuit with direct connection to sensor in potentiostatPotentiostat
Vout
R
C
W
RTIA
ADC
DAC
DOUT
DIN
Readout Circuit
Potentiostat
R
C
W
DACDIN
Proposed ASIC
Current-sensing VCO-based SDM
185Hz
UART Package
DOUT
ISSCC Forum 69 of 78
© 2021 IEEE International Solid-State Circuits Conference
CTDSM Topology
Current-sensing VCO-based 2nd-order noise shaping
ISSCC Forum 70 of 78
© 2021 IEEE International Solid-State Circuits Conference
Dout
sens
or
Cin
Gm
Vbias
Vc
H(s) CCO
CCO
DAC
Iin
I dac
Dual-CCO
Intrinsic CLA
Transfer functions & Coefficients in each stage
Proposed Architecture
∆𝑉= 𝐼𝑠𝐶 = 𝑘𝑠 = 𝑎 𝑓𝑠
∆𝜑 𝑠∆𝑉 = 2𝜋𝐾𝑠 = 𝑘𝑠 = 𝑎 𝑓𝑠→ 𝒂𝟏 = 𝑰𝒊𝒏𝑪𝒊𝒏 · 𝒇𝒔
→ 𝒂𝟐 = 𝟐𝝅𝑲𝑽𝒇𝒔
First-stage
Second-stage
71 of 78ISSCC Forum
© 2021 IEEE International Solid-State Circuits Conference
Circuit Realization
Irefn DiDib
Irefp
15 unit NMOS
cells
I
7.5I
Passive integrator + Dual-CCO structure
Current-steering DAC
15-stage Ring CCO 15-stage Ring CCO
1515
Vin Vbias
Degeneration R-DAC
XOR
DFF Thermometer-to-Binary Adder
15
16
fs
II15
Iin
7.5I
Decoder
ID
R+ R-
ᶵ15
4
4-bit DOUT
Cin If
72 of 78ISSCC Forum
CCO: current controlledring oscillator
© 2021 IEEE International Solid-State Circuits Conference
ASIC Microphotography and Comparison Table
ISSCC Forum 74 of 78
Power Eff.: Input Range/Maximum Consumed Current
© 2021 IEEE International Solid-State Circuits Conference
Measurement Results and Experimental Setup
ISSCC Forum 75 of 78
© 2021 IEEE International Solid-State Circuits Conference
Conclusion
The low-power SOC for out-of/implantable body sensor
networks are introduced
The required challenged circuits, including analog front-end
circuits, passive/active RF front-end circuits, power
management, and digital processing circuits, are presented
A smart application platform to evaluate cardiovascular status of
patients based on the outcomes in clinical research is developed
ISSCC Forum 76 of 78
© 2021 IEEE International Solid-State Circuits Conference
Include Key References Hao-Yun Lee, et. al., “A Power-Efficient Current Readout Circuit with VCO-Based 2nd-Order CT Δ∑ ADC for
Electrochemistry Acquisition,” Accepted by 2020 IEEE A-SSCC, Nov. 2020. Ju-Yi Chen, et. al., “Urine N-terminal pro b-type natriuretic peptide is predictive of heart-failure–related emergency
department visits,” ESC Heart Failure, 02 July 2020, DOI: 10.1002/ehf2.12856. Ding-Siang Ciou, et. al.,, “Colorimetric and Amperometric Detection of Urine Creatinine Based on the ABTS Radical
Cation Modified Electrode,” Sensors & Actuators: B. Chemical vol. 314, July 2020, 128034. Shuenn-Yuh Lee, et. al., “A 2.4 GHz ISM Band OOK Transceiver with High Energy Efficiency for Biomedical Implantable
Applications” IEEE TBioCAS, Feb. 2020. Shuenn-Yuh Lee, et. al., “Electrocardiogram and Phonocardiogram Monitoring System for Cardiac Auscultation,” IEEE
Trans. on Biomedical Circuits and Systems, Dec. 2019. Shuenn-Yuh Lee, et. al., “Development of an Arrhythmia Monitoring System and Human Study,” IEEE Trans. on
Consumer Electronics, Nov. 2018 Shuenn-Yuh Lee, et. al., “Low-Power Wireless ECG Acquisition and Classification System for Body Sensor Networks”,
IEEE Journal of Biomedical and Health Informatics. Jan. 2015. Shuenn-Yuh Lee, et. al., “A Wireless Front-end with Power Management for an Implantable Cardiac Micro-Stimulator,”
IEEE Trans. on Biomedical Circuits and Systems, Feb. 2012. Shuenn-Yuh Lee, et. al., “A Programmable Implantable Micro-stimulator SoC with Wireless Telemetry: Application in
Closed-Loop Endocardial Stimulation for Cardiac Pacemaker,” IEEE TBioCAS, Dec. 2011.
ISSCC Forum 77 of 78
© 2021 IEEE International Solid-State Circuits Conference
Acknowledgement Animal study: thanks for the support of Prof. Y. Y. Chen, National Yang-Ming University,
Taiwan
Thanks all of the students in the CBIC Lab and Doctor J. W. Lin as the acting chairman of the Cardiovascular Center of National Taiwan University Hospital Yun-Lin Branch
Thanks all of co-operated professors (Doctor Ju-Yi Chen) in NCKU hospital
Thanks for Your Attentions !ISSCC Forum 78 of 78